On Thu, 30 Jul 2020 17:17:53 -0700, Don Y
<blockedofcourse@foo.invalid> wrote:
WVR no longer offers this brand and model, it seems. But they ought
to have repair information. I\'d call them up on the phone and ask.
Also, see if they will tell you who made the unit for VWR.
Mine is a Tripp-Lite unit, also 250 watts. I paid $100 in 2006 for a
new unit at the local electronics parts store.
They also make a 500-watt unit. I bet the same outfit makes these for
both brands - only the label varies.
You can also cob together a optically isolated HV probe with a
resistor string driving the LED (with anti-parallel diode protecting
against reverse voltage) in an optocoupler (with at least 2500 V
withstand), and measure the photocurent from the coupler photo diode
or transistor. This will allow you to see if the voltage is remotely
plausible.
Joe Gwinn
<blockedofcourse@foo.invalid> wrote:
On 7/30/2020 4:53 PM, Joe Gwinn wrote:
On Wed, 29 Jul 2020 17:30:51 -0700, Don Y
blockedofcourse@foo.invalid> wrote:
I was hoping to avoid probing the transducers directly. I figured I
could isolate the HV and just verify that the *drive* to it was
working properly. I.e., assume transformer and chokes are intact
so if driven correctly but still no apparent output power, the
transducers must be toast.
I don\'t think that not probing the transduced is a successful
approach. This is a power circuit - not too fussy. Just cob togther
a voltage divider.
I figure if the primary is being chopped, a lack of output power is
either:
- bad xformer
- bad chokes (all 3!)
- bad xducers (all 3)
In either case, I won\'t dick with it any further. (these sorts of things
turn up fairly regularly so its easier just to wait for another to come along)
The frequency will be 30 KHz to 50 KHz, most likely. Won\'t be MHz at
that power level, I\'d guess.
What is the make and model of your unit, anyway?
https://www.ebay.com/itm/VWR-Model-150D-Aquasonic-Ultrasonic-Bath-/143178745430
Apparently it has a different \"part number\" than \"model number\"
so chasing it down on the VWR web site was tedious.
WVR no longer offers this brand and model, it seems. But they ought
to have repair information. I\'d call them up on the phone and ask.
Also, see if they will tell you who made the unit for VWR.
There was no AC power transformer, so an isolation transformer is
essential for working on such a unit. Then, you can ground whatever
part of the circuit is convenient.
There are at least two xformers (IIRC) in the unit. But, one is undoubtedly
for the controls (MCU/display) and the other likely for the transducer step-up.
ISTM that if the entire unit was behind a transformer, that would likely
be pretty (physically) large -- i.e., I should remember having seen
something like that.
Exactly.
Displays blink, buttons seem to be recognized... just no \"bzzzzz\". So, I
assume the controls are working (with possible exception of the interface
to the HV) and the problem lies in the HV section.
I think I have a 500VA isolation transformer lying around from the days
of designing KWHr meters. That\'s probably about the rated size of this load
(4A @ 120VAC).
I have such a transformer. Tripp-Lite? One must open it up and
disconnect the green ground (and isolate the spade terminal with
plastic tubing).
\"Triad-Utrad\". And, apparently only 250VA.
Mine is a Tripp-Lite unit, also 250 watts. I paid $100 in 2006 for a
new unit at the local electronics parts store.
They also make a 500-watt unit. I bet the same outfit makes these for
both brands - only the label varies.
The trick will be accessing the electronics without having to do
a major teardown (the insides look like they are pieced together
from lots of \"modules\" instead of designed as an integrated whole)
But many more points can be accessed, if only you knew what all those
wires were.
Yes, but it\'s a mess in there. Like it was pieced together instead of
being designed as an integrated unit. Like it was designed by someone
in *Detroit* (motor city)!
\"Control processor\" is mounted behind the indicators -- indicators mounted
on the solder side of the (thru hole) board. \"Power\" board mounted on
the base -- with a case-mounted heatsink supporting the switching Q\'s
jutting out over much of the electronics. I.e., if I want to probe stuff,
I will have to remove the boards while staying close to the xducers due to
lead lengths as well as the Qs on the heat sink.
And, of course, ensuring that there\'s water in the tub in case the
thing actually tries to start!
My plan is to trace out the HV section and see how it is gated on/off.
Then, verify the presence of that gate when commanded by the controls.
Then, go looking at the chopper to see if there\'s any signs of life, there.
You can also cob together a optically isolated HV probe with a
resistor string driving the LED (with anti-parallel diode protecting
against reverse voltage) in an optocoupler (with at least 2500 V
withstand), and measure the photocurent from the coupler photo diode
or transistor. This will allow you to see if the voltage is remotely
plausible.
Joe Gwinn