P
Phil Allison
Guest
"John Fields"
** But you could NOT predict the cap value !!!!!!
http://img178.imageshack.us/img178/4863/eico710dipperator.png
That * carefully chosen on test * cap is *series resonating* with the
primary inductance of the particular tranny and has double the AC voltage
across it.
This voltage will go dangerously high when the Oscillator switch ( S1) is
open - using a 275VAC rated class X2 mains cap will be essential. A DC
rated cap will fail short and kill the tranny.
The internal DC voltage is gonna be well down too, scope the secondary
voltage and see why.
Massive asymmetry
Bad idea.
.... Phil
** But that is exactly what it does.There's another certainly less expensive alternative, and that's to
use a capacitor in series with the primary of the dipper's power
transformer in order to use its reactance to drop ~ half the mains
voltage across itself - with very little loss - and put the rest
across the primary of the transformer.
If you're willing to measure, and post, the current into the various
dippers when operating at their rated input voltages, I'll be happy to
show you how to figure out the value of the capacitor.
** The load current for those dippers is gonna be nothing like a sine
wave.
Firstly, there is half wave rectifier in the DC supply - so the iron core
of each AC transformer will be off set creating a highly asymmetrical
current wave.
---
Doesn't matter, as long as it doesn't drive the core into saturation.
---
** But Oz *is* a 50 Hz environment..Then there is the issue of 50Hz operation in lieu of 60Hz - driving the
cores hard into saturation. Normally this produces a spiky current wave
peaking around each zero crossing.
---
That shouldn't be a problem since He's in Oz.
---
The combination of the above currents will look just horrible on a scope.
Putting a series cap in the primary circuit will not behave predictably.
---
I've got an EICO 710 with 1.2ľF in series with its power transformer's
primary, and it works great!
---
** But you could NOT predict the cap value !!!!!!
http://img178.imageshack.us/img178/4863/eico710dipperator.png
That * carefully chosen on test * cap is *series resonating* with the
primary inductance of the particular tranny and has double the AC voltage
across it.
This voltage will go dangerously high when the Oscillator switch ( S1) is
open - using a 275VAC rated class X2 mains cap will be essential. A DC
rated cap will fail short and kill the tranny.
The internal DC voltage is gonna be well down too, scope the secondary
voltage and see why.
Massive asymmetry
Bad idea.
.... Phil