mandag den 30. januar 2023 kl. 00.20.44 UTC+1 skrev John Walliker:
the opamps are only rated for a total supply of 5.5V which it about where the transistor would clamp
so that wouldn\'t do much
On Sunday, 29 January 2023 at 20:11:31 UTC, Fred Bloggs wrote:
That\'s a really weird circuit.
I\'m pretty sure the BAVs are there for ESD protection. Now the hunt is on for the rationale behind BC850 clamp.
The BC850 clamps in conjunction with the 22k input resistors are almost certainly there
to protect the patient from excessive dc under single fault conditions. Direct current
flowing through electrodes attached to the skin can cause severe burns if it is sustained
for a long time - which could easily happen in some medical situations.
Nobody would put 22k resistors in series with the input of a low-noise amplifier if
they didn\'t need to. Their purpose is to limit the current that can flow when driven by the clamping
voltage. Without the clamps the fault current through the electrodes would be around
10 times greater if, for example, one of the BAV99s developed a short to Vcc.
the opamps are only rated for a total supply of 5.5V which it about where the transistor would clamp
so that wouldn\'t do much