On Tue, 12 Aug 2003 14:27:21 +0100, Leigh W3NLB wrote:
10dBm is about normal for that age of analyzer, but they're not usually
usable up to that, the figure to look for is the input level for 1dB
compression. Above that, the analyzer wil start generating significant
spurii of its own. As a rule of thumb, put another 10dB attenuation in,
if everything drops by 10dB, what you see is what you've got. Anything
that drops by more than 10dB is probably an artifact of the instrument.
--
Then there's duct tape ...
(Garrison Keillor)
nofr@sbhevre.pbzchyvax.pb.hx
That's the absolute maximum safe input.My 8554B says "Maximum 10 dBm" on the front in big red letters.
There's probably a reason why they put that on the panel.
10dBm is about normal for that age of analyzer, but they're not usually
usable up to that, the figure to look for is the input level for 1dB
compression. Above that, the analyzer wil start generating significant
spurii of its own. As a rule of thumb, put another 10dB attenuation in,
if everything drops by 10dB, what you see is what you've got. Anything
that drops by more than 10dB is probably an artifact of the instrument.
--
Then there's duct tape ...
(Garrison Keillor)
nofr@sbhevre.pbzchyvax.pb.hx