Isolation transformer source

"Dave Plowman "Pommy Nutcase"

Phil Allison
"JW"

Can anyone recommend a good isolation transformer for sale on-line?

** What kind ??

There are two:

1. Galvanic isolation - for use on a service bench.

2. Neutral conductor isolation.

Most you see offered for sale in the USA are of the *second* kind.


Surely an isolation transformer is just that...

** Shame it is not in the USA - you steaming great fuckwit.



.... Phil
 
In article <87fshcF1r7U1@mid.individual.net>,
Phil Allison <phil_a@tpg.com.au> wrote:
Surely an isolation transformer is just that...

** Shame it is not in the USA - you steaming great fuckwit.
Given your supreme command of the English language it's not a surprise you
don't understand what 'isolate' means...

--
*I don't suffer from insanity -- I'm a carrier

Dave Plowman dave@davenoise.co.uk London SW
To e-mail, change noise into sound.
 
On Thu, 10 Jun 2010 13:42:19 -0700 PlainBill47@yahoo.com wrote in Message
id: <7dj216h43dpi27vfrr5ccb2p11clin2is8@4ax.com>:

On Thu, 10 Jun 2010 09:44:12 -0400, JW <none@dev.null> wrote:

Can anyone recommend a good isolation transformer for sale on-line?
Need one that will handle 5 amps or so at 120VAC, and have a standard AC
plug at the outlet. Prefer one for less than $100 or so, unless that seems
unreasonable. Found one at Mouser for $114 but that's only good for 1.25A.
Getting lots of irrelevant hits on google...

Thanks.
Here's one on eBay. 4.35A for about $75 delivered.
http://cgi.ebay.com/Stancor-GIS-500-Isolation-Transformer-115v-out-115v-/250643065589?cmd=ViewItem&pt=LH_DefaultDomain_0&hash=item3a5b7daaf5
Nice price. I'm working on a 1KW HP 6032A power supply (with outputs
unloaded in this case, so I don't need *too* much current capability) and
need to connect my scope to the hot side of a switching transformer -
specifically the source and drains on the output FETs.

Service manual: (3MB)
http://www.home.agilent.com/agilent/redirector.jspx?action=ref&cname=AGILENT_EDITORIAL&ckey=1000000384-1%3Aepsg%3Aman&lc=eng&cc=US&nfr=-35687.384731
Page 113 shows the schematic of the FET board.

With an isolation transformer, I *think* I would connect the scopes GND
lead the source of Q1 and probe the it's gate to check the it, then move
the scopes GND lead to Q3's drain lead to check it's gate drive.

Given that, would that Ebay transformer be the right choice, and do I need
to isolate (disconnect) the ground on the 6032A power supply under repair?

Thanks.
 
On Fri, 11 Jun 2010 10:30:56 +1000 "Phil Allison" <phil_a@tpg.com.au>
wrote in Message id: <87dee7Fdj9U1@mid.individual.net>:

"JW"

Can anyone recommend a good isolation transformer for sale on-line?

** What kind ??

There are two:

1. Galvanic isolation - for use on a service bench.
Galvanic. Thanks - didn't realize there are different kinds. Don't want to
kill myself!
 
In article <jqp6161lis3v93582a2p4q6g9212udvqpt@4ax.com>,
JW <none@dev.null> wrote:
On Fri, 11 Jun 2010 10:30:56 +1000 "Phil Allison" <phil_a@tpg.com.au
wrote in Message id: <87dee7Fdj9U1@mid.individual.net>:


"JW"

Can anyone recommend a good isolation transformer for sale on-line?

** What kind ??

There are two:

1. Galvanic isolation - for use on a service bench.

Galvanic. Thanks - didn't realize there are different kinds. Don't want
to kill myself!
It certainly seems to me stupid to have the same name for different
devices like this. What is the point of an isolating transformer which
doesn't isolate?

--
*When a clock is hungry it goes back four seconds*

Dave Plowman dave@davenoise.co.uk London SW
To e-mail, change noise into sound.
 
On Fri, 11 Jun 2010 08:24:36 -0500 Chuck <ch@deja.net> wrote in Message
id: <r7e416tt40mloqh91evegbjo9eb2qllqjk@4ax.com>:

On Thu, 10 Jun 2010 09:44:12 -0400, JW <none@dev.null> wrote:

Can anyone recommend a good isolation transformer for sale on-line?
Need one that will handle 5 amps or so at 120VAC, and have a standard AC
plug at the outlet. Prefer one for less than $100 or so, unless that seems
unreasonable. Found one at Mouser for $114 but that's only good for 1.25A.
Getting lots of irrelevant hits on google...

Thanks.


MCM Electronics has one. Chuck
Thanks Chuck. Looking at these for the moment:
http://www.mcmelectronics.com/product/TRIPP-LITE--POWER-PROTECTION--IS500-/28-10160
http://www.mcmelectronics.com/product/TENMA-72-1097-/72-1097

A bit more than I wanted to spend and a little less current capability,
but those may be the breaks.
 
On Sat, 12 Jun 2010 06:53:47 -0400, JW <none@dev.null> wrote:

...
Nice price. I'm working on a 1KW HP 6032A power supply (with outputs
unloaded in this case, so I don't need *too* much current capability) and
need to connect my scope to the hot side of a switching transformer -
specifically the source and drains on the output FETs.

Service manual: (3MB)
http://www.home.agilent.com/agilent/redirector.jspx?action=ref&cname=AGILENT_EDITORIAL&ckey=1000000384-1%3Aepsg%3Aman&lc=eng&cc=US&nfr=-35687.384731
Page 113 shows the schematic of the FET board.

With an isolation transformer, I *think* I would connect the scopes GND
lead the source of Q1 and probe the it's gate to check the it, then move
the scopes GND lead to Q3's drain lead to check it's gate drive.

Given that, would that Ebay transformer be the right choice, and do I need
to isolate (disconnect) the ground on the 6032A power supply under repair?

Thanks.
You need to make 100% sure your life insurance is fully paid up. <g>
 
"JW" <none@dev.null> wrote in message
news:0qo616dhg4nsgshqjsia9l39venjef4bt6@4ax.com...
Nice price. I'm working on a 1KW HP 6032A power supply (with outputs
unloaded in this case, so I don't need *too* much current capability) and
need to connect my scope to the hot side of a switching transformer -
specifically the source and drains on the output FETs.

Service manual: (3MB)
http://www.home.agilent.com/agilent/redirector.jspx?action=ref&cname=AGILENT_EDITORIAL&ckey=1000000384-1%3Aepsg%3Aman&lc=eng&cc=US&nfr=-35687.384731
Page 113 shows the schematic of the FET board.

With an isolation transformer, I *think* I would connect the scopes GND
lead the source of Q1 and probe the it's gate to check the it, then move
the scopes GND lead to Q3's drain lead to check it's gate drive.

Given that, would that Ebay transformer be the right choice, and do I need
to isolate (disconnect) the ground on the 6032A power supply under repair?

Thanks.
This is a "bad idea". Better to use a differential input and two probes. Do
you have a two channel scope with invert on channel 2?

Tom
 
"Dave Plowman " Rabid Pommy Nutcase"


Can anyone recommend a good isolation transformer for sale on-line?

** What kind ??

There are two:

1. Galvanic isolation - for use on a service bench.

2. Neutral conductor isolation.

Most you see offered for sale in the USA are of the *second* kind.


Surely an isolation transformer is just that...

** Shame it is not so in the USA

- you Steaming Great Fuckwit ...............




.... Phil
 
On Sat, 12 Jun 2010 06:53:47 -0400, JW <none@dev.null> wrote:

On Thu, 10 Jun 2010 13:42:19 -0700 PlainBill47@yahoo.com wrote in Message
id: <7dj216h43dpi27vfrr5ccb2p11clin2is8@4ax.com>:

On Thu, 10 Jun 2010 09:44:12 -0400, JW <none@dev.null> wrote:

Can anyone recommend a good isolation transformer for sale on-line?
Need one that will handle 5 amps or so at 120VAC, and have a standard AC
plug at the outlet. Prefer one for less than $100 or so, unless that seems
unreasonable. Found one at Mouser for $114 but that's only good for 1.25A.
Getting lots of irrelevant hits on google...

Thanks.
Here's one on eBay. 4.35A for about $75 delivered.
http://cgi.ebay.com/Stancor-GIS-500-Isolation-Transformer-115v-out-115v-/250643065589?cmd=ViewItem&pt=LH_DefaultDomain_0&hash=item3a5b7daaf5

Nice price. I'm working on a 1KW HP 6032A power supply (with outputs
unloaded in this case, so I don't need *too* much current capability) and
need to connect my scope to the hot side of a switching transformer -
specifically the source and drains on the output FETs.

Service manual: (3MB)
http://www.home.agilent.com/agilent/redirector.jspx?action=ref&cname=AGILENT_EDITORIAL&ckey=1000000384-1%3Aepsg%3Aman&lc=eng&cc=US&nfr=-35687.384731
Page 113 shows the schematic of the FET board.

With an isolation transformer, I *think* I would connect the scopes GND
lead the source of Q1 and probe the it's gate to check the it, then move
the scopes GND lead to Q3's drain lead to check it's gate drive.

Given that, would that Ebay transformer be the right choice, and do I need
to isolate (disconnect) the ground on the 6032A power supply under repair?

Thanks.
That SHOULD be the correct isolation transformer. And yes, the ground
must be isolated. Note that this may well leave the metal chassis of
the power supply at some potentially lethal voltage, so use extreme
caution.

Somewhere in my kit (probably with the isolation transformer I can't
find) is an adapter to use a grounded (three prong) plug in an
ungrounded (two prong) outlet. I used that when working with any
devide that required the isolation transformer. I also carefully
verified that the isolation was indeed effective; I had no desire to
watch my scope probe's ground lead vaporize.

PlainBill
 
In article <hv06fk$jkb$1@news-01.bur.connect.com.au>,
Phil Allison <phil_a@tpg.com.au> wrote:
Surely an isolation transformer is just that...

** Shame it is not so in the USA

- you Steaming Great Fuckwit ...............
But I'm not in the USA and neither are you. However, at least I know not
to use bad language when visiting. If you'd ever known your parents they'd
have likely taught you the same.

--
*If at first you don't succeed, redefine success.

Dave Plowman dave@davenoise.co.uk London SW
To e-mail, change noise into sound.
 
On Sat, 12 Jun 2010 10:32:04 -0400, "tm" <noone@msc.com> wrote:

"JW" <none@dev.null> wrote in message
news:0qo616dhg4nsgshqjsia9l39venjef4bt6@4ax.com...

Nice price. I'm working on a 1KW HP 6032A power supply (with outputs
unloaded in this case, so I don't need *too* much current capability) and
need to connect my scope to the hot side of a switching transformer -
specifically the source and drains on the output FETs.

Service manual: (3MB)
http://www.home.agilent.com/agilent/redirector.jspx?action=ref&cname=AGILENT_EDITORIAL&ckey=1000000384-1%3Aepsg%3Aman&lc=eng&cc=US&nfr=-35687.384731
Page 113 shows the schematic of the FET board.

With an isolation transformer, I *think* I would connect the scopes GND
lead the source of Q1 and probe the it's gate to check the it, then move
the scopes GND lead to Q3's drain lead to check it's gate drive.

Given that, would that Ebay transformer be the right choice, and do I need
to isolate (disconnect) the ground on the 6032A power supply under repair?

Thanks.

This is a "bad idea". Better to use a differential input and two probes. Do
you have a two channel scope with invert on channel 2?

Tom


That would be news to large numbers of electronics technicians who had
to use an isolation transformer when servicing TVs and other 'live
chassis' consumer electronics. There IS a reason the advice was 'to
always keep one hand in your pocket'.

PlainBill
 
In article <j7g716l13t96u9322qmqjiv8r5ppua3tdj@4ax.com>,
<PlainBill47@yahoo.com> wrote:
That would be news to large numbers of electronics technicians who had
to use an isolation transformer when servicing TVs and other 'live
chassis' consumer electronics. There IS a reason the advice was 'to
always keep one hand in your pocket'.
If using a proper isolating transformer, there will be no reference to
ground even with a 'live' chassis. The only way to get an electrical shock
is to touch both outputs from the transformer.

--
*Why isn't there a special name for the back of your knee?

Dave Plowman dave@davenoise.co.uk London SW
To e-mail, change noise into sound.
 
<PlainBill47@yahoo.com> wrote in message
news:j7g716l13t96u9322qmqjiv8r5ppua3tdj@4ax.com...
On Sat, 12 Jun 2010 10:32:04 -0400, "tm" <noone@msc.com> wrote:


"JW" <none@dev.null> wrote in message
news:0qo616dhg4nsgshqjsia9l39venjef4bt6@4ax.com...

Nice price. I'm working on a 1KW HP 6032A power supply (with outputs
unloaded in this case, so I don't need *too* much current capability)
and
need to connect my scope to the hot side of a switching transformer -
specifically the source and drains on the output FETs.

Service manual: (3MB)
http://www.home.agilent.com/agilent/redirector.jspx?action=ref&cname=AGILENT_EDITORIAL&ckey=1000000384-1%3Aepsg%3Aman&lc=eng&cc=US&nfr=-35687.384731
Page 113 shows the schematic of the FET board.

With an isolation transformer, I *think* I would connect the scopes GND
lead the source of Q1 and probe the it's gate to check the it, then move
the scopes GND lead to Q3's drain lead to check it's gate drive.

Given that, would that Ebay transformer be the right choice, and do I
need
to isolate (disconnect) the ground on the 6032A power supply under
repair?

Thanks.

This is a "bad idea". Better to use a differential input and two probes.
Do
you have a two channel scope with invert on channel 2?

Tom


That would be news to large numbers of electronics technicians who had
to use an isolation transformer when servicing TVs and other 'live
chassis' consumer electronics. There IS a reason the advice was 'to
always keep one hand in your pocket'.

PlainBill
You can do it anyway you want. Just that many good scopes have a way to add
the inverted second channel and can be used to look at a signal that is
raised above
ground. No isolation transformer is needed and it is IMO, safer.


YMMV.
 
On Sat, 12 Jun 2010 10:20:10 -0700, PlainBill47 ǝʇoɹʍ:

On Sat, 12 Jun 2010 10:32:04 -0400, "tm" <noone@msc.com> wrote:


"JW" <none@dev.null> wrote in message
news:0qo616dhg4nsgshqjsia9l39venjef4bt6@4ax.com...

Nice price. I'm working on a 1KW HP 6032A power supply (with outputs
unloaded in this case, so I don't need *too* much current capability)
and need to connect my scope to the hot side of a switching
transformer - specifically the source and drains on the output FETs.

Service manual: (3MB)
http://www.home.agilent.com/agilent/redirector.jspx?
action=ref&cname=AGILENT_EDITORIAL&ckey=1000000384-1%3Aepsg%
3Aman&lc=eng&cc=US&nfr=-35687.384731
Page 113 shows the schematic of the FET board.

With an isolation transformer, I *think* I would connect the scopes
GND lead the source of Q1 and probe the it's gate to check the it,
then move the scopes GND lead to Q3's drain lead to check it's gate
drive.

Given that, would that Ebay transformer be the right choice, and do I
need to isolate (disconnect) the ground on the 6032A power supply
under repair?

Thanks.

This is a "bad idea". Better to use a differential input and two
probes. Do you have a two channel scope with invert on channel 2?

Tom


That would be news to large numbers of electronics technicians who had
to use an isolation transformer when servicing TVs and other 'live
chassis' consumer electronics. There IS a reason the advice was 'to
always keep one hand in your pocket'.

PlainBill
Serviced many GE VIR portables with a Sencore PR57 Powerite.
 
Apparently you assume that I didn't understand that.

--
Cheers,
WB
..............


"Phil Allison" <phil_a@tpg.com.au> wrote in message
news:huukoc$q2e$1@news-01.bur.connect.com.au...
The question is not about the number but the nature of the item.


.... Phil
 
PlainBill47@yahoo.com wrote:
On Sat, 12 Jun 2010 10:32:04 -0400, "tm" <noone@msc.com> wrote:


"JW" <none@dev.null> wrote in message
news:0qo616dhg4nsgshqjsia9l39venjef4bt6@4ax.com...

Nice price. I'm working on a 1KW HP 6032A power supply (with outputs
unloaded in this case, so I don't need *too* much current capability) and
need to connect my scope to the hot side of a switching transformer -
specifically the source and drains on the output FETs.

Service manual: (3MB)
http://www.home.agilent.com/agilent/redirector.jspx?action=ref&cname=AGILENT_EDITORIAL&ckey=1000000384-1%3Aepsg%3Aman&lc=eng&cc=US&nfr=-35687.384731
Page 113 shows the schematic of the FET board.

With an isolation transformer, I *think* I would connect the scopes GND
lead the source of Q1 and probe the it's gate to check the it, then move
the scopes GND lead to Q3's drain lead to check it's gate drive.

Given that, would that Ebay transformer be the right choice, and do I need
to isolate (disconnect) the ground on the 6032A power supply under repair?

Thanks.

This is a "bad idea". Better to use a differential input and two probes. Do
you have a two channel scope with invert on channel 2?

Tom


That would be news to large numbers of electronics technicians who had
to use an isolation transformer when servicing TVs and other 'live
chassis' consumer electronics. There IS a reason the advice was 'to
always keep one hand in your pocket'.

That advice was quite valid when two channel scopes weren't seen
outside of a factory or research center.

--
Anyone wanting to run for any political office in the US should have to
have a DD214, and a honorable discharge.
 
"Dave Plowman (News)" wrote:
In article <j7g716l13t96u9322qmqjiv8r5ppua3tdj@4ax.com>,
PlainBill47@yahoo.com> wrote:
That would be news to large numbers of electronics technicians who had
to use an isolation transformer when servicing TVs and other 'live
chassis' consumer electronics. There IS a reason the advice was 'to
always keep one hand in your pocket'.

If using a proper isolating transformer, there will be no reference to
ground even with a 'live' chassis. The only way to get an electrical shock
is to touch both outputs from the transformer.

Not true. You connect the chassis of the UUT to the test equipment,
so ANY voltage you contact inside the UUT is referenced to ground.


--
Anyone wanting to run for any political office in the US should have to
have a DD214, and a honorable discharge.
 
"tm"

You can do it anyway you want. Just that many good scopes have a way to
add
the inverted second channel and can be used to look at a signal that is
raised above ground.
** The problem with that idea is the very limited common mode range
available with that scope function.

The buffers on most ( analogue) scopes become overloaded with input signals
that exceed full screen hight by a factor of 2 or 3 times.

Say you want to examine a 1 volt p-p signal and you set the scope input
atten to 0.5 volts per division - then the largest undistorted input
allowable is gonna be about 0.5 x 8 x 3 = 12 volts p-p.

Not much use with the ACTUAL common mode signal is several hundred volts p-p
from rectified AC supply.


No isolation transformer is needed and it is IMO, safer.
** Fraid that is just not true when dealing with off-line SMPS.



..... Phil
 