T
The Phantom
Guest
On Sat, 06 Mar 2010 15:25:02 -0800, Fred Abse <excretatauris@invalid.invalid>
wrote:
The CT2593-1 described here:
http://www.caltestelectronics.com/www/Cat1GetSubCategory.asp?PN=General%20Purpose%20Differential%20Probes&ID=1.c.1&subcat=Differential%20Probes
is specified for 86 dB CMRR at 50 Hz, 66 dB at 20 kHz, and 1400 volt common mode
rated for only $331
wrote:
On Sat, 06 Mar 2010 16:05:28 -0500, Paul E. Schoen wrote:
"Fred Abse" <excretatauris@invalid.invalid> wrote in message
newsan.2010.03.06.11.18.10.148083@invalid.invalid...
On Fri, 05 Mar 2010 19:05:54 -0800, George Herold wrote:
Why do they make 'scope inputs ground referenced to the third wire of
the AC plug? Today I smoked the 10 ohm resistor that was separating
The AC ground from the 'ground' of my circuit. I'd forgotten to float
the 'scope and connected -15V to the ground clip of the scope probe.
Why not invert one channel and switch to "add".
Then use both inputs differentially.
Most, if not all worthwhile instruments should do that.
Probe compensation needs to be accurate for best CMRR.
Or buy a differential probe?
This will not work for reading a voltage of 1 or 2 volts or less, which is
floating as much as 400 volts above ground. For instance, reading the gate
voltage of an SCR on a 480 VAC mains system. You must float the scope, and
I've done it safely and effectively, with a hand-held scope. There may be
some differential probes that can handle this; I don't know.
The OP was talking about 15 volts common mode, where the technique will
work perfectly well.
56 dB CMR with a common mode voltage of 480 RMS is beyond the
capability of even a 7A13. There are differential probes that will do it,
but they have 4 figure price tags.
The CT2593-1 described here:
http://www.caltestelectronics.com/www/Cat1GetSubCategory.asp?PN=General%20Purpose%20Differential%20Probes&ID=1.c.1&subcat=Differential%20Probes
is specified for 86 dB CMRR at 50 Hz, 66 dB at 20 kHz, and 1400 volt common mode
rated for only $331
I've never needed to go to that extreme. Most of what I do is IGBTs with a
DC bus of around 280V, and the (now occasional) SCR bridge at 200V RMS. A
7A13 will handle those, as will a fairly cheap differential probe (few
hundred bucks).
I hope the handheld you used was battery powered. I wouldn't like to stick
670-something volts peak across the PSU isolation of a floating
line-powered 'scope.