In sci.electronics.equipment Commander Kinsey <CFKinsey@military.org.jp> wrote:
The separate amps jacks exist for at least two reasons:
1) If the range switch also switched in/out the current shunt, then the
user could accidentally create a short circult across the probe tips
simply by turning the range switch to or across the amps measurement
ranges. Even if they were quickly turning /across/ the amps settings
ranges, a brief short circuit would be created, which would cause
damage to the meter as well as the device under test depending upon
what was connected at the time.
2) By having the amps jacks separate, the range switch itself does not
have to have contacts beefy enough to carry the current for the amps
settings. Remember, when measuring amps, the current being measured
flows *through* the meter itself. This would require very different
(and likely much more expensive) range switch contacts.
Some meters do have sensors for the amps jacks that sound an alarm if
the probes are inserted in the amps jacks but the range switch is on a
different setting than amperes.
On Fri, 03 Jul 2020 02:20:53 +0100, RheillyPhoull <Rheilly@bigslong.com> wrote:
On 2/07/2020 10:41 pm, Ralph Mowery wrote:
For a real scare you should see some of the safety movies that
Fluke put out. They show under test conditions what can hapen to
inexpensive meters and their meters under different conditions like
having the meter set for amps and putting across a 480 volt circuit
that has plenty of amps .
Ahh the old \"Leaving it on amps\" trick. How many of us can say they
never did it ?
I find it crazy that you can select volts and have the wires in the
amps holes. The switch should change the contacts. I\'ve broken a
meter doing that, just measuring the voltage on a car battery. £100
meter, but UNFUSED FFS! At least the mA range was fused, but they
couldn\'t be bothered putting in a 20A fuse for the big range....
The separate amps jacks exist for at least two reasons:
1) If the range switch also switched in/out the current shunt, then the
user could accidentally create a short circult across the probe tips
simply by turning the range switch to or across the amps measurement
ranges. Even if they were quickly turning /across/ the amps settings
ranges, a brief short circuit would be created, which would cause
damage to the meter as well as the device under test depending upon
what was connected at the time.
2) By having the amps jacks separate, the range switch itself does not
have to have contacts beefy enough to carry the current for the amps
settings. Remember, when measuring amps, the current being measured
flows *through* the meter itself. This would require very different
(and likely much more expensive) range switch contacts.
Some meters do have sensors for the amps jacks that sound an alarm if
the probes are inserted in the amps jacks but the range switch is on a
different setting than amperes.