I use a HP 5345B and a HP 5370A for time interval measuring ( separate
as well as common mode ).
The 5345A has a jitter stability ( useable resolution ) of 2 nS and
the 5370A a jitter stability of 20 pS.
Both the instruments behave OK.
However, I see that the stability of the period is far better than the
accuracy of the time interval.
Whatever tweaking I have done and whatever signal source I use ( e.g.
a e-11 frequency reference )and at whatever frequency I perform the
measurements, any time interval will be no more stable than about 5e5.
Thus, measuring ( 5370A )( common TI mode ) a 10 Mhz frequency
reference signal the stability shows about +/- 40 pS jitter which is
very good and far within the specs.
Measuring at 1 Mhz ( reference locked ) however, the stability is +/-
some hundreds of pS.
And measuring at 10 Khz , the stability is +/- 50 nS
I get about the same ( relative ) results when using the 5345B instead
of the 5370A ( be it that the jitter stability and resolution of the
5345B is 100 x less ).
However, measuring the period ( instead of the TI mode ) I get the
expected results of about e-11( which is cf the specs if locked to a
better than e-11 freq.ref ).
In the frequency modes both the instruments perfom far within specs (
stability resolution better than e-11 ).
What is it that TI measurement ( in these particular HP gear )is so
much less stable than period measurent ?
Since both the instruments show this very phenomenum and for both even
in the same relationship at varying frequencies, I guess this must be
system feature and no flaw.
Any HP guru in this group ?
I bent my mind for days but I cannot resolve this one.
Thanks for help.
Peter, PE1E ex PE1ECM
as well as common mode ).
The 5345A has a jitter stability ( useable resolution ) of 2 nS and
the 5370A a jitter stability of 20 pS.
Both the instruments behave OK.
However, I see that the stability of the period is far better than the
accuracy of the time interval.
Whatever tweaking I have done and whatever signal source I use ( e.g.
a e-11 frequency reference )and at whatever frequency I perform the
measurements, any time interval will be no more stable than about 5e5.
Thus, measuring ( 5370A )( common TI mode ) a 10 Mhz frequency
reference signal the stability shows about +/- 40 pS jitter which is
very good and far within the specs.
Measuring at 1 Mhz ( reference locked ) however, the stability is +/-
some hundreds of pS.
And measuring at 10 Khz , the stability is +/- 50 nS
I get about the same ( relative ) results when using the 5345B instead
of the 5370A ( be it that the jitter stability and resolution of the
5345B is 100 x less ).
However, measuring the period ( instead of the TI mode ) I get the
expected results of about e-11( which is cf the specs if locked to a
better than e-11 freq.ref ).
In the frequency modes both the instruments perfom far within specs (
stability resolution better than e-11 ).
What is it that TI measurement ( in these particular HP gear )is so
much less stable than period measurent ?
Since both the instruments show this very phenomenum and for both even
in the same relationship at varying frequencies, I guess this must be
system feature and no flaw.
Any HP guru in this group ?
I bent my mind for days but I cannot resolve this one.
Thanks for help.
Peter, PE1E ex PE1ECM