Hi all,
I've completed my tests of the main transformer and am now 99% certain
that it is the cause of all the problems I've been experiencing with this
old analogue scope. It's clear there's something very wrong with the
large, multi-tapped output winding. Here's the schematic again:
https://www.flickr.com/photos/128859641@N02/24535280896/in/dateposted-
I removed ALL connections from the transformer. ALL the other output
windings are giving exactly the outputs I would expect from a given
input; it's just the long winding on the lower right hand side that's
giving nonsense outputs. As you can see, the centre tap is grounded and
there are 3 tapping points either side of it. When injected with a 20kHz
sine wave of 50V p-p to the primary winding, the peak-to-peak outputs
from the problem secondary at each tap are as follows (from top to bottom)
13V
13V
3V
0V (gnd)
3V
1.8V
1.8V
I would have expected these voltages to be symmetrical either side of the
0V centre tap, but as you can see, this isn't the case at all. I can only
conclude from this, to use a technical term, that this tranny is fucked.
If there's something obvious I've overlooked (which I doubt) please feel
free to point it out. Otherwise I'll be opening it up to perform an
autopsy over the weekend.
Thanks again to everyone who tried to help.
I've completed my tests of the main transformer and am now 99% certain
that it is the cause of all the problems I've been experiencing with this
old analogue scope. It's clear there's something very wrong with the
large, multi-tapped output winding. Here's the schematic again:
https://www.flickr.com/photos/128859641@N02/24535280896/in/dateposted-
I removed ALL connections from the transformer. ALL the other output
windings are giving exactly the outputs I would expect from a given
input; it's just the long winding on the lower right hand side that's
giving nonsense outputs. As you can see, the centre tap is grounded and
there are 3 tapping points either side of it. When injected with a 20kHz
sine wave of 50V p-p to the primary winding, the peak-to-peak outputs
from the problem secondary at each tap are as follows (from top to bottom)
13V
13V
3V
0V (gnd)
3V
1.8V
1.8V
I would have expected these voltages to be symmetrical either side of the
0V centre tap, but as you can see, this isn't the case at all. I can only
conclude from this, to use a technical term, that this tranny is fucked.
If there's something obvious I've overlooked (which I doubt) please feel
free to point it out. Otherwise I'll be opening it up to perform an
autopsy over the weekend.
Thanks again to everyone who tried to help.