i recently purchased a dead Tek 465b scope (no lights except for power and
uncal, not even scale illumination) and opened it up to find no voltages
near spec at the test points on the main board. resistance checks showed
the +55v test point near ground. disconnecting the connector to the board
next to the crt (670-6385-?) allowed all voltages to come up. reconnecting
this board and disconnecting the coiled timing delay line that comes in from
the vertical preamp board to this board also allows +55v to come up. I
thought the delay line might be shorted out, and reconnected it to this
board, and disconnected the other end (in the vertical preamp board). Same
problem with the +55v.
I put an ohmmeter across the terminals of the delay line at both ends, and
there is no evidence of a short. yet with the delay line plugged into the
670-6385 board the +55v is loaded down, and thus the entire power supply
output is drawn down.
I'm confused. it seems that this coiled timing delay line is causing
problems, but it shows no sign of shorting or damage. am i missing
something? (besides a good service manual and schematics, which i am trying
to obtain) what is the likelihood that this line has failed?
and what does this board do? it has inputs from the vertical preamp board,
also power, sep. enable, and beam find thru the power connector, and outputs
to the crt tube.
uncal, not even scale illumination) and opened it up to find no voltages
near spec at the test points on the main board. resistance checks showed
the +55v test point near ground. disconnecting the connector to the board
next to the crt (670-6385-?) allowed all voltages to come up. reconnecting
this board and disconnecting the coiled timing delay line that comes in from
the vertical preamp board to this board also allows +55v to come up. I
thought the delay line might be shorted out, and reconnected it to this
board, and disconnected the other end (in the vertical preamp board). Same
problem with the +55v.
I put an ohmmeter across the terminals of the delay line at both ends, and
there is no evidence of a short. yet with the delay line plugged into the
670-6385 board the +55v is loaded down, and thus the entire power supply
output is drawn down.
I'm confused. it seems that this coiled timing delay line is causing
problems, but it shows no sign of shorting or damage. am i missing
something? (besides a good service manual and schematics, which i am trying
to obtain) what is the likelihood that this line has failed?
and what does this board do? it has inputs from the vertical preamp board,
also power, sep. enable, and beam find thru the power connector, and outputs
to the crt tube.