We brought a couple of new electronic designers onto the ventilator project I\'m working on. One of them was suggesting we control the motor with a variable voltage rather than an H-bridge to switch the current in a PWM manner. He mentioned some high power amps to drive this, but they are class AB and so would have high losses at anything other than the maximum output.
Is it a common practice to use a switching circuit to supply a controlled voltage to a motor? The motor is an inherently inductive load, so a PWM drive at a high enough frequency would approximate a constant current to the motor. A smoothing cap at the source (we already have 2000 uF) helps that a lot. The H-bridge would be retained to allow the polarity to the motor to be reversed.
Is this a method used? It would lower the I2R heating in the motor with a lower continuous current rather than a higher pulsed current with a duty cycle.
We have scope plots showing a 500 Hz pulse rate is not fast enough to make that work. I need to ask if we can get some better measurements to see how long it takes for the motor current to drop significantly when the H-bridge removes the power connection and the reverse diode conducts.
The H-bridge has a 10 kHz max rate, but that might be good enough. I\'ll see if I can get someone to make a measurement.
--
Rick C.
- Get 1,000 miles of free Supercharging
- Tesla referral code - https://ts.la/richard11209
Is it a common practice to use a switching circuit to supply a controlled voltage to a motor? The motor is an inherently inductive load, so a PWM drive at a high enough frequency would approximate a constant current to the motor. A smoothing cap at the source (we already have 2000 uF) helps that a lot. The H-bridge would be retained to allow the polarity to the motor to be reversed.
Is this a method used? It would lower the I2R heating in the motor with a lower continuous current rather than a higher pulsed current with a duty cycle.
We have scope plots showing a 500 Hz pulse rate is not fast enough to make that work. I need to ask if we can get some better measurements to see how long it takes for the motor current to drop significantly when the H-bridge removes the power connection and the reverse diode conducts.
The H-bridge has a 10 kHz max rate, but that might be good enough. I\'ll see if I can get someone to make a measurement.
--
Rick C.