Spectra Physics 120 HeNe / 256 Exciter problem.

A

Andrew Back

Guest
Hello,

I recently picked up an old Spectra Physics 120 HeNe laser head plus 256 Exciter, and I'm not getting any laser output. However, the tube glow looks to be around the right sort of colour and I don't think there is air leakage..

I've yet to try cleaning the optics, but I have checked the tube voltage and current and I'm not sure if I have a problem here, as it's 2.7kV and 6mA, whereas Sam's Laser FAQ suggests it should be 3.7kV and 7mA. Also, after being powered up for a while the power supply seems to drop out -- I get a buzzing sound and the plasma goes intermittent, before stopping altogether. When it's intermittent the tube voltage spikes up to around 4-5kV (at least that's what I see it jump to on my DMM). So, perhaps a PSU component is on its way out?

I'd be grateful of any suggestions that folks may have. I'm hoping it's simply a case of resistor or cap that has not aged well.

Cheers,

Andrew
 
Andrew Back wrote:

Hello,

I recently picked up an old Spectra Physics 120 HeNe laser head plus 256
Exciter, and I'm not getting any laser output. However, the tube glow
looks to be around the right sort of colour and I don't think there is air
leakage.

I've yet to try cleaning the optics, but I have checked the tube voltage
and current and I'm not sure if I have a problem here, as it's 2.7kV and
6mA, whereas Sam's Laser FAQ suggests it should be 3.7kV and 7mA. Also,
after being powered up for a while the power supply seems to drop out -- I
get a buzzing sound and the plasma goes intermittent, before stopping
altogether. When it's intermittent the tube voltage spikes up to around
4-5kV (at least that's what I see it jump to on my DMM). So, perhaps a PSU
component is on its way out?
How old is it? Most glass-case HeNe lasers slowly leak the Helium
away. You still get the pinkish glow from the He-Ne mixture, but there
isn't enough He left to develop the population inversion of excited states
to produce laser action. if the tube is over 20 years old, that is the most
likely cause.

Are the mirrors separate from the HeNe tube? If so, this is an "open
cavity" laser, and the mirrors need VERY careful alignment to get laser
action. If the mirrors are glued to the end of the tube, then no alignment
is possible.


Jon
 
Andrew Back <arback@gmail.com> writes:

Hello,

I recently picked up an old Spectra Physics 120 HeNe laser head plus
256 Exciter, and I'm not getting any laser output. However, the tube
glow looks to be around the right sort of colour and I don't think
there is air leakage.

I've yet to try cleaning the optics, but I have checked the tube
voltage and current and I'm not sure if I have a problem here, as
it's 2.7kV and 6mA, whereas Sam's Laser FAQ suggests it should be
3.7kV and 7mA. Also, after being powered up for a while the power
supply seems to drop out -- I get a buzzing sound and the plasma
goes intermittent, before stopping altogether. When it's
intermittent the tube voltage spikes up to around 4-5kV (at least
that's what I see it jump to on my DMM). So, perhaps a PSU component
is on its way out?

I'd be grateful of any suggestions that folks may have. I'm hoping it's
simply a case of resistor or cap that has not aged well.

Hi:

The tube probably IS at least one of the problems assuming no one has
messed with the mirror alignment.

There's a range of current that would be OK. I'd suggest increasing it
to 7 mA. That should stop it from dropping out. I don't think there
is anything wrong electrically but the tube has leaked. If you can run
it for a few hours or days, it may recover at least partially.

--
sam | Sci.Electronics.Repair FAQ: http://www.repairfaq.org/
Repair | Main Table of Contents: http://www.repairfaq.org/REPAIR/
+Lasers | Sam's Laser FAQ: http://www.repairfaq.org/sam/lasersam.htm
| Mirror Sites: http://www.repairfaq.org/REPAIR/F_mirror.html

Important: Anything sent to the email address in the message header above is
ignored unless my full name AND either lasers or electronics is included in the
subject line. Or, you can contact me via the Feedback Form in the FAQs.
 
Jon Elson <jmelson@wustl.edu> writes:

Andrew Back wrote:

Hello,

I recently picked up an old Spectra Physics 120 HeNe laser head plus 256
Exciter, and I'm not getting any laser output. However, the tube glow
looks to be around the right sort of colour and I don't think there is air
leakage.

I've yet to try cleaning the optics, but I have checked the tube voltage
and current and I'm not sure if I have a problem here, as it's 2.7kV and
6mA, whereas Sam's Laser FAQ suggests it should be 3.7kV and 7mA. Also,
after being powered up for a while the power supply seems to drop out -- I
get a buzzing sound and the plasma goes intermittent, before stopping
altogether. When it's intermittent the tube voltage spikes up to around
4-5kV (at least that's what I see it jump to on my DMM). So, perhaps a PSU
component is on its way out?
How old is it? Most glass-case HeNe lasers slowly leak the Helium
away. You still get the pinkish glow from the He-Ne mixture, but there
isn't enough He left to develop the population inversion of excited states
to produce laser action. if the tube is over 20 years old, that is the most
likely cause.

Are the mirrors separate from the HeNe tube? If so, this is an "open
cavity" laser, and the mirrors need VERY careful alignment to get laser
action. If the mirrors are glued to the end of the tube, then no alignment
is possible.

This ia an SP-120, probably from the 1970s. And, yes, those old tubes
do leak due to the Epoxy seals. Modern tubes do not leak.

The dropout current is higher than 6 mA, so he should increase the current
to 7 mA and let it run. It may recover, at least partially.

If no one has attempted to adjust alignment or clean the optics, they
are probably OK. Else, all bets are off. It could be that someone found
it didn't lase and really mucked it up. :( :)

--
sam | Sci.Electronics.Repair FAQ: http://www.repairfaq.org/
Repair | Main Table of Contents: http://www.repairfaq.org/REPAIR/
+Lasers | Sam's Laser FAQ: http://www.repairfaq.org/sam/lasersam.htm
| Mirror Sites: http://www.repairfaq.org/REPAIR/F_mirror.html

Important: Anything sent to the email address in the message header above is
ignored unless my full name AND either lasers or electronics is included in the
subject line. Or, you can contact me via the Feedback Form in the FAQs.
 
Hi Sam,

On Wednesday, 5 March 2014 21:16:34 UTC, Samuel M. Goldwasser wrote:

The tube probably IS at least one of the problems assuming no one has

messed with the mirror alignment.

Ah, so I'll look into sourcing a bottle of helium.

There's a range of current that would be OK. I'd suggest increasing it

to 7 mA. That should stop it from dropping out. I don't think there

is anything wrong electrically but the tube has leaked. If you can run

it for a few hours or days, it may recover at least partially.

I've turned the current up to 7mA and will see how it gets on. Any idea why the voltage is only ~2.7kV and what I could try to get the exciter output back up to 3.7kV?

Looking into the end of the tube the optics do look grubby, but I get the impression that attempting to clean them may not be helpful.

Thanks for the advice, it's much appreciated!

Regards,

Andrew
 
It still dropped out with the current set at 7mA. After a while a buzzing sound develops which becomes increasingly low frequency until the plasma is lost, and sometimes it will then strike back up again. Not sure why it's dropped out unless some component is heating up and causing failure -- perhaps related to the voltage being down also.

Cheers,

Andrew

On Thursday, 6 March 2014 17:54:55 UTC, Andrew Back wrote:
Hi Sam,



On Wednesday, 5 March 2014 21:16:34 UTC, Samuel M. Goldwasser wrote:



The tube probably IS at least one of the problems assuming no one has



messed with the mirror alignment.



Ah, so I'll look into sourcing a bottle of helium.



There's a range of current that would be OK. I'd suggest increasing it



to 7 mA. That should stop it from dropping out. I don't think there



is anything wrong electrically but the tube has leaked. If you can run



it for a few hours or days, it may recover at least partially.



I've turned the current up to 7mA and will see how it gets on. Any idea why the voltage is only ~2.7kV and what I could try to get the exciter output back up to 3.7kV?



Looking into the end of the tube the optics do look grubby, but I get the impression that attempting to clean them may not be helpful.



Thanks for the advice, it's much appreciated!



Regards,



Andrew
 
On 03/05/2014 01:19 PM, Samuel M. Goldwasser wrote:
Jon Elson <jmelson@wustl.edu> writes:

Andrew Back wrote:

Hello,

I recently picked up an old Spectra Physics 120 HeNe laser head plus 256
Exciter, and I'm not getting any laser output. However, the tube glow
looks to be around the right sort of colour and I don't think there is air
leakage.

I've yet to try cleaning the optics, but I have checked the tube voltage
and current and I'm not sure if I have a problem here, as it's 2.7kV and
6mA, whereas Sam's Laser FAQ suggests it should be 3.7kV and 7mA. Also,
after being powered up for a while the power supply seems to drop out -- I
get a buzzing sound and the plasma goes intermittent, before stopping
altogether. When it's intermittent the tube voltage spikes up to around
4-5kV (at least that's what I see it jump to on my DMM). So, perhaps a PSU
component is on its way out?
How old is it? Most glass-case HeNe lasers slowly leak the Helium
away. You still get the pinkish glow from the He-Ne mixture, but there
isn't enough He left to develop the population inversion of excited states
to produce laser action. if the tube is over 20 years old, that is the most
likely cause.

Are the mirrors separate from the HeNe tube? If so, this is an "open
cavity" laser, and the mirrors need VERY careful alignment to get laser
action. If the mirrors are glued to the end of the tube, then no alignment
is possible.

This ia an SP-120, probably from the 1970s. And, yes, those old tubes
do leak due to the Epoxy seals. Modern tubes do not leak.

The dropout current is higher than 6 mA, so he should increase the current
to 7 mA and let it run. It may recover, at least partially.

If no one has attempted to adjust alignment or clean the optics, they
are probably OK. Else, all bets are off. It could be that someone found
it didn't lase and really mucked it up. :( :)

You actually repaired SPs? We were a dealer and they wanted to do all
the repairs themselves. Remember the 3600 rotator? Now you can get a
rotatinglaser level for way under a hundred bucks! I don't miss the
sewer business, don't get me wrong!
 
On Thursday, 6 March 2014 21:39:29 UTC, Samuel M. Goldwasser wrote:

It still dropped out with the current set at 7mA. After a while a

buzzing sound develops which becomes increasingly low frequency

until the plasma is lost, and sometimes it will then strike back up

again. Not sure why it's dropped out unless some component is

heating up and causing failure -- perhaps related to the voltage

being down also.



If it's low even when the discharge is stable, then likely tube contamination.

Interesting. Aside from stopping lasing, what does air contamination do in terms of plasma stability / tube voltage?

I doubt a He soak will help. Unfortunately, it's not uncommon for SP-120s to

have tubes that won't stay lit even at 8 mA though the color looks good.

Ordered the He now, but I am sure it will find other uses and I'll at least give it a go.

One test to confirm it's the tube and not the power supply is to put a fan

on the tube. If it's the tube, that should enable it to stay lit at least

for a longer time.

So air contamination means it heats up and when it gets too hot the plasma drops out?

When I first got it (only a few weeks ago) it stayed lit for 4 or 5 hours. Whereas now it drops out after being powered up for 30-45 minutes.

(This assumes that the 3.7 kV value in the Laser FAQ is correct. It's from

the spec sheet but I have not measured a healthy one.)

Right, good to know.

> Possibly time to install a modern laser tube or head. :)

To be honest I only picked this up because I like the design of older gas lasers and what might be regarded as classic tubes. Oh well.

Thanks again.

Andrew
 
dave <ricketzz@earthlink.net> writes:

On 03/05/2014 01:19 PM, Samuel M. Goldwasser wrote:
Jon Elson <jmelson@wustl.edu> writes:

Andrew Back wrote:

Hello,

I recently picked up an old Spectra Physics 120 HeNe laser head plus 256
Exciter, and I'm not getting any laser output. However, the tube glow
looks to be around the right sort of colour and I don't think there is air
leakage.

I've yet to try cleaning the optics, but I have checked the tube voltage
and current and I'm not sure if I have a problem here, as it's 2.7kV and
6mA, whereas Sam's Laser FAQ suggests it should be 3.7kV and 7mA. Also,
after being powered up for a while the power supply seems to drop out -- I
get a buzzing sound and the plasma goes intermittent, before stopping
altogether. When it's intermittent the tube voltage spikes up to around
4-5kV (at least that's what I see it jump to on my DMM). So, perhaps a PSU
component is on its way out?
How old is it? Most glass-case HeNe lasers slowly leak the Helium
away. You still get the pinkish glow from the He-Ne mixture, but there
isn't enough He left to develop the population inversion of excited states
to produce laser action. if the tube is over 20 years old, that is the most
likely cause.

Are the mirrors separate from the HeNe tube? If so, this is an "open
cavity" laser, and the mirrors need VERY careful alignment to get laser
action. If the mirrors are glued to the end of the tube, then no alignment
is possible.

This ia an SP-120, probably from the 1970s. And, yes, those old tubes
do leak due to the Epoxy seals. Modern tubes do not leak.

The dropout current is higher than 6 mA, so he should increase the current
to 7 mA and let it run. It may recover, at least partially.

If no one has attempted to adjust alignment or clean the optics, they
are probably OK. Else, all bets are off. It could be that someone found
it didn't lase and really mucked it up. :( :)


You actually repaired SPs? We were a dealer and they wanted to do all
the repairs themselves. Remember the 3600 rotator? Now you can get a
rotatinglaser level for way under a hundred bucks! I don't miss the
sewer business, don't get me wrong!

These are not laser level type lasers. ;-)

--
sam | Sci.Electronics.Repair FAQ: http://www.repairfaq.org/
Repair | Main Table of Contents: http://www.repairfaq.org/REPAIR/
+Lasers | Sam's Laser FAQ: http://www.repairfaq.org/sam/lasersam.htm
| Mirror Sites: http://www.repairfaq.org/REPAIR/F_mirror.html

Important: Anything sent to the email address in the message header above is
ignored unless my full name AND either lasers or electronics is included in the
subject line. Or, you can contact me via the Feedback Form in the FAQs.
 
Andrew Back <arback@gmail.com> writes:

It still dropped out with the current set at 7mA. After a while a
buzzing sound develops which becomes increasingly low frequency
until the plasma is lost, and sometimes it will then strike back up
again. Not sure why it's dropped out unless some component is
heating up and causing failure -- perhaps related to the voltage
being down also.

If it's low even when the discharge is stable, then likely tube contamination.
I doubt a He soak will help. Unfortunately, it's not uncommon for SP-120s to
have tubes that won't stay lit even at 8 mA though the color looks good.

One test to confirm it's the tube and not the power supply is to put a fan
on the tube. If it's the tube, that should enable it to stay lit at least
for a longer time.

(This assumes that the 3.7 kV value in the Laser FAQ is correct. It's from
the spec sheet but I have not measured a healthy one.)

Possibly time to install a modern laser tube or head. :)

--
sam | Sci.Electronics.Repair FAQ: http://www.repairfaq.org/
Repair | Main Table of Contents: http://www.repairfaq.org/REPAIR/
+Lasers | Sam's Laser FAQ: http://www.repairfaq.org/sam/lasersam.htm
| Mirror Sites: http://www.repairfaq.org/REPAIR/F_mirror.html

Important: Anything sent to the email address in the message header above is
ignored unless my full name AND either lasers or electronics is included in the
subject line. Or, you can contact me via the Feedback Form in the FAQs.
 
On 03/06/2014 01:30 PM, Samuel M. Goldwasser wrote:
dave <ricketzz@earthlink.net> writes:

On 03/05/2014 01:19 PM, Samuel M. Goldwasser wrote:
Jon Elson <jmelson@wustl.edu> writes:

Andrew Back wrote:

Hello,

I recently picked up an old Spectra Physics 120 HeNe laser head plus 256
Exciter, and I'm not getting any laser output. However, the tube glow
looks to be around the right sort of colour and I don't think there is air
leakage.

I've yet to try cleaning the optics, but I have checked the tube voltage
and current and I'm not sure if I have a problem here, as it's 2.7kV and
6mA, whereas Sam's Laser FAQ suggests it should be 3.7kV and 7mA. Also,
after being powered up for a while the power supply seems to drop out -- I
get a buzzing sound and the plasma goes intermittent, before stopping
altogether. When it's intermittent the tube voltage spikes up to around
4-5kV (at least that's what I see it jump to on my DMM). So, perhaps a PSU
component is on its way out?
How old is it? Most glass-case HeNe lasers slowly leak the Helium
away. You still get the pinkish glow from the He-Ne mixture, but there
isn't enough He left to develop the population inversion of excited states
to produce laser action. if the tube is over 20 years old, that is the most
likely cause.

Are the mirrors separate from the HeNe tube? If so, this is an "open
cavity" laser, and the mirrors need VERY careful alignment to get laser
action. If the mirrors are glued to the end of the tube, then no alignment
is possible.

This ia an SP-120, probably from the 1970s. And, yes, those old tubes
do leak due to the Epoxy seals. Modern tubes do not leak.

The dropout current is higher than 6 mA, so he should increase the current
to 7 mA and let it run. It may recover, at least partially.

If no one has attempted to adjust alignment or clean the optics, they
are probably OK. Else, all bets are off. It could be that someone found
it didn't lase and really mucked it up. :( :)


You actually repaired SPs? We were a dealer and they wanted to do all
the repairs themselves. Remember the 3600 rotator? Now you can get a
rotatinglaser level for way under a hundred bucks! I don't miss the
sewer business, don't get me wrong!

These are not laser level type lasers. ;-)

They aren't self leveling but otherwise they work great for drop
ceilings and poured floors. SP had a cool alignment range (box of mirrors).
 
"Samuel M. Goldwasser" <sam@repairfaq.org> wrote in message
news:u61nrdn1q.fsf@repairfaq.org...
Andrew Back <arback@gmail.com> writes:

It still dropped out with the current set at 7mA. After a while a
buzzing sound develops which becomes increasingly low frequency
until the plasma is lost, and sometimes it will then strike back up
again. Not sure why it's dropped out unless some component is
heating up and causing failure -- perhaps related to the voltage
being down also.

If it's low even when the discharge is stable, then likely tube
contamination.
I doubt a He soak will help. Unfortunately, it's not uncommon for SP-120s
to
have tubes that won't stay lit even at 8 mA though the color looks good.

One test to confirm it's the tube and not the power supply is to put a fan
on the tube. If it's the tube, that should enable it to stay lit at least
for a longer time.

(This assumes that the 3.7 kV value in the Laser FAQ is correct. It's
from
the spec sheet but I have not measured a healthy one.)

Possibly time to install a modern laser tube or head. :)

--

Don't these power supplies have a several watt carbon comp ballast resistor?
They are well known to drift up in value and/or become unstable when heating
up.

Just a guess.

tm
 
Samuel M. Goldwasser wrote:


This ia an SP-120, probably from the 1970s. And, yes, those old tubes
do leak due to the Epoxy seals. Modern tubes do not leak.
I hate to dispute a known expert in the field, but as you say
this laser may be almost FORTY years old! Yes, the crummy old tubes with
sloppy epoxy seals may have leaked down in a few years, but even the
best HeNe all-glass tubes will eventually lose their He. It diffuses
right through the glass. And, the partially aluminum tubes
leaked much faster, still.

Jon
 
On Friday, 7 March 2014 04:32:23 UTC, Jon Elson wrote:
Samuel M. Goldwasser wrote:





This ia an SP-120, probably from the 1970s. And, yes, those old tubes

do leak due to the Epoxy seals. Modern tubes do not leak.

I hate to dispute a known expert in the field, but as you say

this laser may be almost FORTY years old! Yes, the crummy old tubes with

sloppy epoxy seals may have leaked down in a few years, but even the

best HeNe all-glass tubes will eventually lose their He. It diffuses

right through the glass. And, the partially aluminum tubes

leaked much faster, still.

Sam said earlier that it _will_ have lost some He and so that will be at least one of the problems it has. The question then was whether it has air contamination also.

I'm going to try giving the a helium soak, but if it came to having to break the seal, evacuate it, bake out and refill -- well, that's probably beyond me.

Regards,

Andrew
 
On Thursday, 6 March 2014 21:39:29 UTC, Samuel M. Goldwasser wrote:

One test to confirm it's the tube and not the power supply is to put a fan

on the tube. If it's the tube, that should enable it to stay lit at least

for a longer time.

I tried this and it didn't seem to have an effect. A write up of experiences so far, along with a few photos, can be found at:

http://designspark.com/eng/blog/breathing-life-into-a-vintage-gas-laser

(This assumes that the 3.7 kV value in the Laser FAQ is correct. It's from

the spec sheet but I have not measured a healthy one.)

This still has me wondering and in particular when I look at those old resistors and caps etc...

Helium ordered so will try a soak in any case.

> Possibly time to install a modern laser tube or head. :)

Never! :eek:)

Cheers,

Andrew
 
On 03/06/2014 08:32 PM, Jon Elson wrote:
Samuel M. Goldwasser wrote:


This ia an SP-120, probably from the 1970s. And, yes, those old tubes
do leak due to the Epoxy seals. Modern tubes do not leak.
I hate to dispute a known expert in the field, but as you say
this laser may be almost FORTY years old! Yes, the crummy old tubes with
sloppy epoxy seals may have leaked down in a few years, but even the
best HeNe all-glass tubes will eventually lose their He. It diffuses
right through the glass. And, the partially aluminum tubes
leaked much faster, still.

Jon

I was told the gases become un-mixed. The Helium and Neon separate when
the tube isn't used. You have to use them weekly to mix the gas, if
nothing else.
 
dave wrote:


I was told the gases become un-mixed. The Helium and Neon separate when
the tube isn't used. You have to use them weekly to mix the gas, if
nothing else.

That's quite humorous! Gases will not spontaneously separate out
by gravity, you can probably find this in a physics textbook. While
theoretically this could happen at VERY low temperatures, it is impossible
at room temperature. I have glass HeNe lasers that are about 20 years
old, and they developed the laser beam instantly when turned on.
I also have some other odd lasers including an external cavity one, that
will not lase no matter how long you run it, and a He soak was attempted
without success.

Jon
 
Jon Elson <elson@pico-systems.com> writes:

Samuel M. Goldwasser wrote:


This ia an SP-120, probably from the 1970s. And, yes, those old tubes
do leak due to the Epoxy seals. Modern tubes do not leak.
I hate to dispute a known expert in the field, but as you say
this laser may be almost FORTY years old! Yes, the crummy old tubes with
sloppy epoxy seals may have leaked down in a few years, but even the
best HeNe all-glass tubes will eventually lose their He. It diffuses
right through the glass. And, the partially aluminum tubes
leaked much faster, still.

Not on any time scale that matters. :) New/MOS HeNes from the early '80s
with hard seals and even before when hard-seals were first introduced will
still work like new. So, yes, everything leaks eventually but they have
pretty much nailed the He diffusion through glass thing. That's not to
say every tube will be like that but it is more common than not.

--
sam | Sci.Electronics.Repair FAQ: http://www.repairfaq.org/
Repair | Main Table of Contents: http://www.repairfaq.org/REPAIR/
+Lasers | Sam's Laser FAQ: http://www.repairfaq.org/sam/lasersam.htm
| Mirror Sites: http://www.repairfaq.org/REPAIR/F_mirror.html

Important: Anything sent to the email address in the message header above is
ignored unless my full name AND either lasers or electronics is included in the
subject line. Or, you can contact me via the Feedback Form in the FAQs.
 
Andrew Back <arback@gmail.com> writes:

On Thursday, 6 March 2014 21:39:29 UTC, Samuel M. Goldwasser wrote:

One test to confirm it's the tube and not the power supply is to put a fan

on the tube. If it's the tube, that should enable it to stay lit at least

for a longer time.

I tried this and it didn't seem to have an effect. A write up of experiences so far, along with a few photos, can be found at:

http://designspark.com/eng/blog/breathing-life-into-a-vintage-gas-laser

(This assumes that the 3.7 kV value in the Laser FAQ is correct. It's from

the spec sheet but I have not measured a healthy one.)

This still has me wondering and in particular when I look at those old resistors and caps etc...

Helium ordered so will try a soak in any case.

Please keep us informed. If you have a spectrometer/monochromator you can
take a look at the discharge spectra and see if there are any O2/N2 linse,
and whether those two yellow He/Ne lines show up and are about equal in
intensity. See the Laser FAQ. ;-)

Possibly time to install a modern laser tube or head. :)

Never! :eek:)

Or a diode laser. ;-)

Ah come on, no one will know. "You're getting 250 mW from that head???". :)

--
sam | Sci.Electronics.Repair FAQ: http://www.repairfaq.org/
Repair | Main Table of Contents: http://www.repairfaq.org/REPAIR/
+Lasers | Sam's Laser FAQ: http://www.repairfaq.org/sam/lasersam.htm
| Mirror Sites: http://www.repairfaq.org/REPAIR/F_mirror.html

Important: Anything sent to the email address in the message header above is
ignored unless my full name AND either lasers or electronics is included in the
subject line. Or, you can contact me via the Feedback Form in the FAQs.
 
dave <ricketzz@earthlink.net> writes:

On 03/06/2014 08:32 PM, Jon Elson wrote:
Samuel M. Goldwasser wrote:


This ia an SP-120, probably from the 1970s. And, yes, those old tubes
do leak due to the Epoxy seals. Modern tubes do not leak.
I hate to dispute a known expert in the field, but as you say
this laser may be almost FORTY years old! Yes, the crummy old tubes with
sloppy epoxy seals may have leaked down in a few years, but even the
best HeNe all-glass tubes will eventually lose their He. It diffuses
right through the glass. And, the partially aluminum tubes
leaked much faster, still.

Jon

I was told the gases become un-mixed. The Helium and Neon separate
when the tube isn't used. You have to use them weekly to mix the gas,
if nothing else.

Balderdahs. ;)

--
sam | Sci.Electronics.Repair FAQ: http://www.repairfaq.org/
Repair | Main Table of Contents: http://www.repairfaq.org/REPAIR/
+Lasers | Sam's Laser FAQ: http://www.repairfaq.org/sam/lasersam.htm
| Mirror Sites: http://www.repairfaq.org/REPAIR/F_mirror.html

Important: Anything sent to the email address in the message header above is
ignored unless my full name AND either lasers or electronics is included in the
subject line. Or, you can contact me via the Feedback Form in the FAQs.
 

Welcome to EDABoard.com