Blurred trace on Gould/Advance OS260

R

Rick

Guest
Can anyone help me repair an old Gould OS260 oscilloscope?

Really, I need a circuit diagram or service manual. This is the first
'scope I have tried to repair, I could really do with a second one so
please help me get it going :)

The trace will not focus, in XY mode the 'dot' is about 5mm square!
The focus control has no effect, neither does the internal focus
preset or any other control. Everything seems to work fine other than
the bad focus problem, which persists on every setting.

I haven't replaced any electrolytic capacitors yet, most of them
(4.7uF 500V) read from 4 to 6 ohms ESR - not sure if this is really
bad or not.

Could it be a gassy tube? I hope not :-(

Any suggestions gratefully accepted - can anyone give me a hint of
where to look for this problem?

One final query - there are four cement-coated resistors that are in
line with some of the leads going to the tube. These are too hot to
touch without some pain - is this normal?

Thanks,
Rick.
 
Rick wrote:

The trace will not focus, in XY mode the 'dot' is about 5mm square!
The focus control has no effect, neither does the internal focus
preset or any other control. Everything seems to work fine other than
the bad focus problem, which persists on every setting.

I haven't replaced any electrolytic capacitors yet, most of them
(4.7uF 500V) read from 4 to 6 ohms ESR - not sure if this is really
bad or not.

One final query - there are four cement-coated resistors that are in
line with some of the leads going to the tube. These are too hot to
touch without some pain - is this normal?
This points to a lot of leakage through the caps! They are usually
the part that dies anyway. Try to identify the caps on the focus
circuit and replace one at a time.
-Dan Barlow
 
On Wed, 29 Oct 2003 14:24:54 -0500, Dan Barlow <danbarlow@radix.net>
wrote:

Rick wrote:

The trace will not focus, in XY mode the 'dot' is about 5mm square!
The focus control has no effect, neither does the internal focus
preset or any other control. Everything seems to work fine other than
the bad focus problem, which persists on every setting.

I haven't replaced any electrolytic capacitors yet, most of them
(4.7uF 500V) read from 4 to 6 ohms ESR - not sure if this is really
bad or not.

One final query - there are four cement-coated resistors that are in
line with some of the leads going to the tube. These are too hot to
touch without some pain - is this normal?

This points to a lot of leakage through the caps! They are usually
the part that dies anyway. Try to identify the caps on the focus
circuit and replace one at a time.
-Dan Barlow

Thanks for that, I'll try tomorrow.

Rick.
 
I've just spotted that some previous owner has removed and not
replaced a few components! A couple I can guesstimate (supply section)
but a couple are in the focus circuit. D'oh!

Now I really need a schematic. There are part numbers on the board but
not values.

Can anyone help?

Rick.
 